求教hp3070问题

以检测输出电压是不是正常。一般的是GPIP线除了问题
如果gpib出了问题通常会boot失败。你们有过gpib线有问题,但能正常boot的时候么?
我们现在又有了新的现象,可能与上面的问题有关:
在测试的过程中,当自检电源时报错:
R350 The power supplies are in over-voltage protection.The over-voltage levels have been programmed to the maximum allowable level for each supply.Check to see that this voltage has not been exeeded on the board.The over-voltage protection circuitvy may also be tripped if small negative voltages (even tenths of volts) are present on the board, or if there is a problem with the power supply sense connections to the board.One supply with an over-voltage condition may cause the other supplied to pull over-voltage,,so be sure to check all supplies at line 2284.
这个和什么可能有关系??
要是我就把板子和电源断开,测试电源的输出。看看有没有短路的地方?然后再检查其他的电源
或者把电源关掉再开一遍试试看。
1.自检电源,看是哪一组输出有问题。
2.你如果确认GPIP没有问题,建议你确认是否对应的电源有问题?
6624种的电源一般也经常坏。
哈哈,问题解决了,有可能是fixture 短路了,反正自己好了。
如果fixture短路的话,因该在short test时就坏了啊,怎么会表现在power 部分啊,奇怪。
R350 - Power Supply in Overvoltage Protection

Date: May 97
Revision: B.02.xx

The "R350 - Power supply in overvoltage protection" error is generated because the system is detecting a discrepancy between the sense and the source voltages at the board under test.

R350 troubleshooting suggestions:

Is your power supply wiring correct?
Check for probes not making contact, stuck probes, etc.
Check for power sense and supply miswiring. Use the fixture/trace file to find the power sense and supply probing locations.
Another test to check your power supply sense wiring: expose the power probes and directly jumper the high side of the supply to the high sense probe and the low side of the supply to the low sense probe. Now, run the test without the fixture.
If you no longer see the R350 error, this indicates that the sense is not correctly detecting the power supply voltage when you are running your fixture.

Check the traces on your board. Do the same test described in the previous paragraph, but run the test with the board and fixture on the testhead. If you no longer see the R350 error, this indicates that there may be a bad trace on the board or your board description is incorrect.
Is your device under test (DUT) power supply bad?
If you are experiencing the R350 error on multiple testheads with the same fixture and board, this is probably not the case.
To check your DUT power supplies:
Turn on the DUT power supplies.
Directly on the back of the power supplies, measure the voltage from signal high to signal low using a voltmeter.
Measure the voltage from sense high to sense low using a voltmeter.
These two measured values (for each supply) need to be identical (within a certain marginal error that is set for overvoltage for each supply. Check your power supply manual for the exact value). If the measured values are not identical for one (or more) of your supplies, this points to a power supply that may be faulty. If you find this to be the case, you can further verify the integrity of the power supplies by running full hardware diagnostics on your system.
Is the topology of your board affecting the order in which the power supplies are brought up?
We have seen the R350 in situations where it was a case of order sensitivity based on the topology of the board. Setting up the power supplies in the opposite order or using the "wait" option on the "sps" command ensures that the supplies come up in the order that you need them to.
Is the correct board on the fixture?
The R350 can be generated when you have the incorrect board on the fixture.
If utilizing parallel power supplies, are they set up correctly?
For the recommended procedure on setting up parallel power supplies, reference the Agilent 3070 SeriesII Users' Manual: Test & Fixture Development - Connecting DUT Supplies in Parallel.
Are your power supplies set-up to provide the power your board requires?
Check to make sure the voltage and current your board requires was correctly described to the system so enough probes are assigned.
这些文件应该可以对你有帮助!
经常有小电容误测是何故?

求教hp3070问题

请问,在软件没有问题的情况下:
导致6624A的输出电压在电路板上测量不到的原因有可能是什么??(电路板已确认没有问题)
这个问题我昨天碰到过,其实你们都错了。
真正的原因是TAp线有问题,老是这样的情况可以试着更改他的CURRENT LIMIT,are you ok?